Illustrated cross-section of a keratoconic eye with a color height map over the cornea and the white of the eye, warmer where the surface sits higher

What is scleral profilometry?

Scroll through how a scan of the front of the eye, from the cornea out onto the white, becomes a scleral lens shaped to land evenly. Then read how reliable it is and how it compares.

By the Scleral Lens Team · Updated October 28, 2026 · 23 published sources cited

Scroll to see each step

  1. 01 Mapping past the cornea

    A scleral lens rests on the white of the eye, beyond the cornea. Profilometry scans both, so the lens can be designed to fit the shape it will land on.

  2. 02 The white isn’t round

    Few scleras are evenly round: most sit higher or lower in some directions. A round lens edge can rest on the high side and lift where the white sits lower.

  3. 03 A drop of dye

    Instruments that read the tear film need a little fluorescein dye to see it; others, using a special camera, don’t. Your lids are held open so more of the white can be measured.

  4. 04 Light patterns, several views

    A pattern of light falls on the eye, and cameras record how it bends over the surface. Nothing touches the eye. You look at a target, and with some instruments up and down too.

  5. 05 Stitched into one surface

    Software joins the views into one 3D surface of the cornea and the white. Where the lids still hid the eye, it estimates the shape, and those estimates are less reliable.

  6. 06 A height map, read by direction

    The map is colored by height. The fitter reads how deep the eye is at the lens’s landing diameter in each direction. Here the white sits lower above than below.

  7. 07 A landing zone shaped to match

    The lens’s landing zone is designed from the map, toric, quadrant-specific or free-form, so it lands evenly all the way round. Your fitter still checks it on the eye, and it may need changes.

Illustration, not to scale. Steps simplified. Your fitter’s process may differ.

The short answer

Scleral profilometry is an imaging scan that measures the height and shape of your cornea and the surrounding white of the eye. Fitters can use the map to design a scleral lens, including a landing zone shaped to match an uneven sclera. Studies so far show it can work, with fitting effort similar to traditional trial-lens fitting, but it doesn't replace a fitter's judgment. Most scleral lenses are still fitted with trial lenses.

Key points

  • A profilometer maps the shape of the cornea and sclera, which an ordinary corneal topographer doesn't reach.
  • Most scleras aren't evenly round: in one series of 140 eyes, only 5.7% were.
  • The map can guide toric, quadrant-specific, or free-form lens designs.
  • Your eyelids can block part of the scan; fitters may need to hold them open.
  • In surveys, image-based fitting is used for a minority of fits, and needs about the same number of lenses as trial-lens fitting.

What it measures

A standard corneal topographer maps the cornea, the clear front window of your eye, not the white around it. A scleral lens, though, rests on the white of the eye, beyond the cornea. A corneo-scleral profilometer extends the map outward, measuring the height and shape of both the cornea and the sclera. Fitters use that information to choose how deep the lens must be to clear your cornea, and how its landing zone should be shaped to sit evenly. Profilometers are also used in research to measure scleral toricity, meaning how much steeper the sclera is in one direction than another.[7]

A 2025 review lists profilometry-based fitting among the advances that help achieve a fit in difficult eyes, such as those with ectasia, a previous corneal transplant, or corneal scarring.[8]

Why the shape of the white of the eye matters

  • Most scleras aren’t evenly round. In 140 eyes measured with a scleral topographer, 5.7% had a mostly round (spherical) sclera and 28.6% were regularly toric. The other 65.7% had raised or lowered areas or irregular patterns that a standard round or toric lens edge isn’t designed for.[9] A review sums it up as only about 6% of scleras being symmetric.[10]
  • The farther out, the less even. In healthy eyes, the front of the eye is nearly evenly round 12.8 mm across but more uneven 15 mm across.[11] Unevenness grows with distance from the center of the cornea.[12]
  • Keratoconus adds to it. Eyes with keratoconus show more corneo-scleral asymmetry than healthy eyes, especially at moderate and advanced stages.[13]
  • What a round lens does on an uneven white. In 10 healthy eyes wearing the same rotationally symmetric scleral lens, the more toric the sclera, the more the lens flexed and the more unevenly it pressed along its main meridians.[14] At the slit lamp, a fitter may see the edge lifting in one direction while it sits well in the opposite one.[9]

How a scan is done, step by step

Profilometers work in different ways, and your fitter’s routine may differ. These are the steps the research describes.

1. Mapping past the cornea

The scan covers the cornea and as much of the white of the eye as the instrument can see. The instruments are cameras and light; they don’t touch the eye, though the lids are held open. Images took a median of 10 minutes to acquire in a survey of practitioners, and about 10 minutes per eye in a clinical study.[2][1]

2. The white isn’t round

The point of measuring past the cornea is the landing zone. Where the white sits lower than a round lens edge expects, the edge can lift; where it sits higher, the edge can press (see above).

3. A drop of dye

Instruments that read the tear film need it to show up. For one fringe-projection profilometer, the Eye Surface Profiler, fluorescein dye is required, in a more viscous solution than saline; in one study the fluorescein strip was moistened with a lubricating eye drop.[17] Instruments based on a Scheimpflug camera, such as the corneo-scleral profile module of the Pentacam, capture images independent of the tear film.[4]

The eyelids are held open, gently, so more of the white can be measured.[4] In one study, holding the lids open increased the measured area by 25% to 35% at the diameters where scleral lenses land.[4]

4. Patterns of light, from several directions

A fringe-projection profilometer shines a pattern of light onto the eye and calculates the surface’s height from how the pattern bends over it.[15] The Eye Surface Profiler takes its picture with you looking straight ahead at a target, in a single shot.[17][16] Other instruments take several views and join them: the sMap3D with you looking up, straight ahead and down,[18] and the Pentacam’s corneo-scleral module looking straight ahead, up, down, left and right.[4]

Why several views? In 25 eyes, a straight-ahead image covered only 39% of the eye’s surface 16 mm across, compared with 93% for three views joined together.[18]

5. Stitched into one surface

Software joins the views into one 3D map of the cornea and the white.[18] Where the lids or lashes still hid the eye, the Pentacam module extends (extrapolates) the measured data to fill the map, and how much was really measured changes those estimates.[4]

6. A height map, read by direction

The result is a height map of the front of the eye.[15] Fitters read the sagittal height (how deep the eye is across a given diameter), its highest and lowest values around the eye, and the difference between them, which is a measure of toricity.[19][5]

7. A landing zone shaped to match

Data from scans or impressions allow lenses to be made that closely align with the white of the eye.[3] The landing zone can be made toric, meaning different in two directions,[7] or the whole back surface can be designed free-form from the scan.[1] Your fitter still checks the lens on your eye, and it may need changes.[1]

Other ways to measure the white of the eye

  • Optical coherence tomography (OCT). Anterior segment OCT can measure sagittal height and the angle of the white in each direction it images; one study measured eight directions.[11] A small pilot study used OCT measurements of corneal and scleral height to design large prosthetic devices.[21]
  • Scheimpflug imaging. The Pentacam’s corneo-scleral module measured a smaller area than the Eye Surface Profiler in one comparison (about 14.8 mm vs 16.4 mm across).[19]

How the map shapes the lens

  • Toric landing zones. In 10 healthy eyes, landing zones with 100 to 200 microns of toricity reduced lens flexure by about 62% and lens rotation by about 77% compared with a round landing zone, without changing how far the lens decentered.[7] The scleral toricity in that study was measured with a profilometer. See what a toric landing zone is.
  • Free-form designs. In a study of eyes with ocular surface disease, lenses were designed free-form from the scan; results are below.[1]
  • Remote design. In one case report, a keratoconus patient with very toric scleras whose earlier fits hadn’t worked was scanned in Ireland, the lens was designed in Italy, and the first lenses worked without adjustment.[23] One case can’t tell you how often that happens.

How reliable the scans are

  • Repeatability. In 25 eyes of people presenting for scleral lens fitting, one profilometer gave repeat measurements of eye depth that agreed within 100 microns in 95% of cases, and the authors concluded it was repeatable enough for fitting.[5] Another profilometer gave consistent depth measurements across diameters of 11 to 15 mm in 35 healthy eyes, with the edge of the cornea and the white less repeatable than the cornea.[20]
  • Accuracy falls toward the edge. Tested on calibrated surfaces, the Eye Surface Profiler’s error was under 10 microns over the central 8 mm and under 40 microns over 16 mm, but reached up to 200 microns at the very edge of its range.[15]
  • Eyelids get in the way. The lids cover part of the sclera. In 20 healthy adults, holding the lids open increased the amount of measured surface substantially at some distances from center, and the instrument’s estimates for hidden areas differed between natural and held-open lid positions, most of all at the top of the eye.[4] Where the lids hide the surface, the software estimates the missing shape, and those estimates depend on how much of the eye was actually measured.[4]
  • One view isn’t enough for some instruments. With half the data removed, the measured toricity changed by half a diopter or more in 28% of eyes.[18]
  • Joining views adds error. In 38 healthy people, maps stitched from several views of a corneal topographer differed from a single-shot profilometer by up to 98 microns, and the authors concluded stitching isn’t yet ready to fully replace single-shot measurement.[16] The Eye Surface Profiler reads a fluorescein-stained tear film, and the quality of the tear film can affect the measurement.[17]
  • Instruments don’t agree with each other. The Pentacam module and the Eye Surface Profiler gave significantly different eye depths and toricity in the same 38 people, so their numbers can’t be swapped.[19] If you change practices, a new scan on the new practice’s instrument is reasonable.
  • Time of day. In 109 healthy eyes scanned five times between morning and evening, the shape of the sclera didn’t change significantly over the day.[6]

What the studies show about fitting

  • Ocular surface disease. In a prospective study of 15 eyes of 9 patients, scans took an average of about 10 minutes per eye. Fitting needed an average of 2.9 follow-up visits and 2.1 lenses. One eye couldn’t be fitted from the scan because of insufficient data, and one didn’t tolerate lens wear. All 13 remaining eyes were fitted successfully. The authors noted that the imaging did not remove the need for skilled clinical observation.[1]
  • Efficiency. In a survey of 423 practitioners who fit scleral lenses for keratoconus, image-based fitting needed about the same first-visit time and number of lenses as trial-lens fitting. Images took a median of 10 minutes to acquire and were adequate to order a first lens a median of 80% of the time.[2]
  • How often it’s used. In the same survey population, practitioners who had image-based technology used it for a median of 5% of their keratoconus fits. Most fits were still done with trial lenses.[3]
  • Vision. In 23 eyes with keratoconus, pellucid marginal degeneration or ectasia after LASIK, profilometry-guided scleral lenses improved average vision from about 20/80 to about 20/20 and reduced higher-order aberrations. There was no comparison group, so this shows the lenses helped, not that a scan beats trial lenses.[22]

No study we found shows that profilometry needs fewer lenses or visits than trial-lens fitting. The studies that compare them found about the same.

How it compares with other ways of fitting

Approach How the shape is found Touches the eye? Read more
Trial-lens fit Lenses of known shape tried on the eye, then adjusted Yes (trial lenses) What a fitting involves
Profilometry (scan-based) An optical scan of the cornea and white, turned into a design No; dye for some instruments This page
Impression-based A soft mold of the front of the eye, scanned in 3D Yes (impression material) Lenses made from an impression
PROSE A treatment program with its own device design process Yes (trial devices) What is PROSE?

In the practitioner survey, image-based and impression-based fitting took about the same first-visit time and number of lenses as trial-lens fitting; first images were adequate to order a lens a median of 80% of the time, and first impressions 90%.[2] Many practitioners with a scanner or impressions still fit most eyes with trial lenses.[3] For help choosing, see impression-based, scan-based, PROSE or standard.

What this means for you

Profilometry is a tool, not a guarantee. A fitter with a profilometer isn’t automatically better than one without; what matters is how well they use whatever methods they have. Scans are especially helpful when your sclera is unevenly shaped, when standard lenses keep landing poorly, or when a toric or quadrant-specific landing zone is being considered. Some fitters capture the eye’s shape with a physical mold instead: see lenses made from an impression of the eye.

If you’re choosing a fitter, it’s reasonable to ask which fitting methods they use and how they’d approach your eye. See questions to ask a scleral lens fitter.

Common questions

Does profilometry mean I won't need trial lenses?

Sometimes the first lens ordered from a scan is the one you keep, but not always. You may still try lenses on and need adjustments. In studies, the number of lenses and visits was similar to traditional fitting.

Is the scan uncomfortable?

It doesn't touch your eye. You look at a target while the instrument takes images, and the fitter may gently hold your eyelids open so more of the white of the eye is captured. Tell them if you need a break.

Why do I have to look up and down during the scan?

Your eyelids cover part of the white of the eye. Some instruments take pictures with you looking straight ahead and then up and down (or side to side), and join them into one map, so areas hidden in one view are caught in another. In one study, a single straight-ahead picture covered well under half of the eye's surface 16 mm across, while three joined views covered almost all of it.

Why did they put yellow dye in my eye?

Some profilometers read light from the tear film and need a little fluorescein dye, the same dye used for many eye exams, to see the surface. Instruments that use a Scheimpflug camera don't need it. Ask your fitter which kind they use.

Is profilometry better than an impression?

They're different tools. A scan doesn't touch the eye but can miss areas hidden by the lids. An impression captures the surface physically, including very irregular areas. Fitters choose based on your eye and what they have available.

Keep reading

What is a toric landing zone?

The landing zone is the outer ring of a scleral lens that rests on the white of the eye. Many eyes are steeper in some directions than others, so a lens with an evenly curved landing zone can press harder in some spots and lift in others. A toric landing zone is curved differently in two directions to match that shape, and quadrant-specific or fully custom landing zones go further. In small studies, toric landing zones reduced lens flexing and rotation and improved tear exchange.

Impression-based scleral lenses: how they're made

An impression-based scleral lens is made from a mold of the front of your eye instead of from a set of trial lenses. A soft silicone impression material sets into a negative of your eye's surface, the mold is scanned in 3D, and a lens is cut on a lathe to match. Its edge can follow an uneven white of the eye, a raised area, or the bump of a glaucoma drainage device. EyePrintPRO is cleared by the FDA as a rigid gas permeable contact lens. It's mostly used when standard scleral lenses haven't fit, and small studies report better vision for many people, along with refits and some complications.

What are mini-scleral lenses?

A mini-scleral lens is a scleral lens that extends only a little past the edge of your cornea. In the classification most often cited, a mini-scleral lens is up to 6 mm wider than the visible colored part of your eye, and a large scleral lens is more than 6 mm wider. Both vault over the cornea and rest on the white of the eye; they just differ in how far they reach. Your fitter picks the size that suits your eye's shape and your condition.

What is scleral lens vault, or clearance?

Vault, also called clearance, is the gap between the back of a scleral lens and the front of your cornea, filled with saline. Too little and the lens can touch the cornea as it settles; too much and less oxygen reaches the cornea and vision can be affected. Published reviews cite targets of roughly 100 to 200 or 100 to 300 microns after the lens settles, but they also stress that the right vault is whatever keeps your vision clear, your eye comfortable, and your cornea healthy.

Impression-based, scan-based, PROSE, or standard: which scleral lens should I try?

They're different ways of designing the same kind of lens. Standard fitting uses trial lenses from a set. Scan-based fitting designs the lens from a map of your eye's surface. Impression-based lenses are made from a mold of your eye. PROSE is a customized scleral lens treatment program from one organization and its partner clinics. Most people do well with standard fitting by an experienced fitter. The more customized routes are usually considered for very irregular eyes, or after standard fits haven't worked, and the fitter's skill matters at least as much as the method.

What a scleral lens fitting involves

A scleral lens fitting is a series of visits, not a single appointment. Your fitter examines and maps your eyes, fits or designs a lens, orders a custom pair, teaches you to put them in and take them out, and then fine-tunes the fit over follow-up visits. In an international survey, an average fit used a little over two lenses per patient across a few visits. Expect it to take weeks rather than days.

0

Sources

  1. Yoon H, Yoon H, Harthan JS, et al. Process and outcomes of fitting corneoscleral profilometry-driven scleral lenses for patients with ocular surface disease. Eye Contact Lens. 2024;50(3):132-137. doi:10.1097/ICL.0000000000001064 pubmed.ncbi.nlm.nih.gov
  2. Shorter E, Fogt J, Nau C, et al. Image- and impression-based technology for scleral lens fitting for keratoconus: efficiency of the fitting process. Cont Lens Anterior Eye. 2024;47(5):102174. doi:10.1016/j.clae.2024.102174 pubmed.ncbi.nlm.nih.gov
  3. Fogt JS, Schornack M, Nau C, et al. Image- and impression-based technology in scleral lens fitting for keratoconus: availability and utilization. Eye Contact Lens. 2024;50(7):292-296. doi:10.1097/ICL.0000000000001100 pubmed.ncbi.nlm.nih.gov
  4. Hughes RPJ, Iqbal A, Hoffmann G, et al. Agreement between extrapolated corneoscleral topographical data obtained during natural and retracted eyelid positions. Ophthalmic Physiol Opt. 2025;45(3):627-636. doi:10.1111/opo.13473 pubmed.ncbi.nlm.nih.gov
  5. DeNaeyer G, Sanders DR. sMap3D corneo-scleral topographer repeatability in scleral lens patients. Eye Contact Lens. 2018;44(Suppl 1):S259-S264. doi:10.1097/ICL.0000000000000401 pubmed.ncbi.nlm.nih.gov
  6. Barberán-Bernardos L, Ariza-Gracia MÁ, Piñero DP. Diurnal variation in corneo-scleral morphology. J Optom. 2026;19(2):100593. doi:10.1016/j.optom.2025.100593 pubmed.ncbi.nlm.nih.gov
  7. Alexander J, Belaineh Aweke Y, Bhebhe Z, et al. The effect of landing zone toricity on scleral lens fitting characteristics and optics. Ophthalmic Physiol Opt. 2024;44(5):867-875. doi:10.1111/opo.13324 pubmed.ncbi.nlm.nih.gov
  8. Gindina S, Kang JJ, Jacobs DS. Scleral lenses for correction of irregular astigmatism: advances and limitations. Curr Opin Ophthalmol. 2025;36(4):282-287. doi:10.1097/ICU.0000000000001149 pubmed.ncbi.nlm.nih.gov
  9. Macedo-de-Araújo RJ, Fadel D, Barnett M. How can we best measure the performance of scleral lenses? Current insights. Clin Optom (Auckl). 2022;14:47-65. doi:10.2147/OPTO.S284632 pmc.ncbi.nlm.nih.gov
  10. Rojas-Viñuela J, Frogozo MJ, Piñero DP. What we know about the scleral profile and its impact on contact lens fitting. Clin Exp Optom. 2023;106(6):591-604. doi:10.1080/08164622.2022.2097858 pubmed.ncbi.nlm.nih.gov
  11. Ritzmann M, Caroline PJ, Börret R, Korszen E. An analysis of anterior scleral shape and its role in the design and fitting of scleral contact lenses. Cont Lens Anterior Eye. 2018;41(2):205-213. doi:10.1016/j.clae.2017.10.010 pubmed.ncbi.nlm.nih.gov
  12. Consejo A, Llorens-Quintana C, Bartuzel MM, Iskander DR, Rozema JJ. Rotation asymmetry of the human sclera. Acta Ophthalmol. 2019;97(2):e266-e270. doi:10.1111/aos.13901 pubmed.ncbi.nlm.nih.gov
  13. Piñero DP, Martínez-Abad A, Soto-Negro R, et al. Differences in corneo-scleral topographic profile between healthy and keratoconus corneas. Cont Lens Anterior Eye. 2019;42(1):75-84. doi:10.1016/j.clae.2018.05.005 pubmed.ncbi.nlm.nih.gov
  14. Eltantawy E. Assessing scleral toricity induced lens flexure and contact disparity using anterior segment optical coherence tomography. Cont Lens Anterior Eye. 2025;48(3):102386. doi:10.1016/j.clae.2025.102386 pubmed.ncbi.nlm.nih.gov
  15. Iskander DR, Wachel P, Simpson PN, Consejo A, Jesus DA. Principles of operation, accuracy and precision of an Eye Surface Profiler. Ophthalmic Physiol Opt. 2016;36(3):266-278. doi:10.1111/opo.12292 pubmed.ncbi.nlm.nih.gov
  16. Lin WP, Wu LY, Lin WR, et al. To stitch or not to stitch, that is the question: multi-gaze eye topography stitching versus single-shot profilometry. Photonics. 2025;12(4):318. doi:10.3390/photonics12040318 doi.org
  17. Consejo A, Wu R, Abass A. Anterior scleral regional variation between Asian and Caucasian populations. J Clin Med. 2020;9(11):3419. doi:10.3390/jcm9113419 pmc.ncbi.nlm.nih.gov
  18. DeNaeyer G, Sanders DR, Farajian TS. Surface coverage with single vs. multiple gaze surface topography to fit scleral lenses. Cont Lens Anterior Eye. 2017;40(3):162-169. doi:10.1016/j.clae.2017.03.009 pubmed.ncbi.nlm.nih.gov
  19. Bandlitz S, Esper P, Stein M, Dautzenberg T, Wolffsohn JS. Corneoscleral topography measured with Fourier-based profilometry and Scheimpflug imaging. Optom Vis Sci. 2020;97(9):766-774. doi:10.1097/OPX.0000000000001572 pubmed.ncbi.nlm.nih.gov
  20. Bataille L, Molina-Martin A, Piñero DP. Intrasession repeatability of corneal, limbal and scleral measurements obtained with a Fourier transform profilometer. Cont Lens Anterior Eye. 2021;44(5):101382. doi:10.1016/j.clae.2020.11.002 pubmed.ncbi.nlm.nih.gov
  21. Le HG, Tang M, Ridges R, Huang D, Jacobs DS. Pilot study for OCT guided design and fit of a prosthetic device for treatment of corneal disease. J Ophthalmol. 2012;2012:812034. doi:10.1155/2012/812034 pubmed.ncbi.nlm.nih.gov
  22. Barone V, Petrini D, Surico PL, et al. Profilometry-guided scleral lenses improve visual acuity and reduce ocular aberrations in irregular corneas: a retrospective case series. Eye Contact Lens. 2025;51(10):445-449. doi:10.1097/ICL.0000000000001199 pubmed.ncbi.nlm.nih.gov
  23. Fadel D, Gildea C. Case report: remote scleral lens fitting for high toric scleras in a keratoconus patient. Optom Vis Sci. 2023;100(12):876-881. doi:10.1097/OPX.0000000000002085 pubmed.ncbi.nlm.nih.gov

Last updated October 28, 2026. Found an error or a newer study? Let us know and we'll correct the page.